When do you use bvm & Nonrebreather?

ThadeusJ

Forum Lieutenant
240
69
28
Mask/circuit connectors must follow an ISO standard (ISO 5356-1) which is also captured in the ISO standard for BVM connectors, ISO 10651-4. The masks must either have a 22mm female connection or a 15mm male connection. The is why most masks have both, with a 22mmOD/15mm ID connector. Infant masks usually have the smaller connector (15mm O.D.).

Some CPAP masks have a swivel/elbow which can get in the way, but the connections should conform to the same standards. Connector bushings may be needed if attaching certain nebulizer tees or other ancillary equipment.
 
Top